when I set the conf register to 0xd5, the ch1 register to 0x05, using internal reference. The signal's amplitude is around 7.5 mV, which is not the same as 2x-(Vrefp-VrefN)/2400 in the datasheet.
I have read this link [常见问题]关于TI ADS129x系列生物势ADC的常见问题 - 数据转换器论坛 - 数据转换器 - TI 端对端支持论坛. However, I did not find the solution. If convenient ,could you offer me a link, thanks!
Our device expert is out-of-office, so please expect some delay. We will respond to this query the week of December 2nd.
Thank you for your post. You are observing the correct signal amplitude. The amplitude is calculated as:
Amplitude (V) = 4.5V / 2400 = 0.001875, or 1.875mV. For 2x setting (CAL_AMP = 1), the amplitude will be 3.75mV, which is the same as 7.5mV peak-to-peak.