bq34z100监测 12V铅蓄电池,chemid获取失败 测试数据以及chemid反馈见附件4428.chemid.zip8321.chemid-report.rar
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
bq34z100监测 12V铅蓄电池,chemid获取失败 测试数据以及chemid反馈见附件4428.chemid.zip8321.chemid-report.rar
建议你把节数设成6。因为12V的铅酸电池内部本来就是6节铅酸电池串联的。
按照您说的把电池改成6节,得到的一个相对好点的chemid 烧录进去后,进行电量自学习时:电池充满后静置vok复位但是upstate一直未04未更新为01,整个过程中电量未自动更新一直显示为零3326.chemid-report.rar