工具与软件:
我将这款 BMS 用于我们的含两节电池的葡萄糖计产品。 根据数据表、它应该会以3A 的电流充电、但根据我的测试、我无法使用该电流为电池充电、而且当我检查 CRG_STAT 寄存器显示为0时、也是如此。 我使用 Arduino Uno 进行开发。 请帮我解决这个问题、对我来说、这是一个瓶颈问题。 我 也可以分享测试代码。
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
工具与软件:
我将这款 BMS 用于我们的含两节电池的葡萄糖计产品。 根据数据表、它应该会以3A 的电流充电、但根据我的测试、我无法使用该电流为电池充电、而且当我检查 CRG_STAT 寄存器显示为0时、也是如此。 我使用 Arduino Uno 进行开发。 请帮我解决这个问题、对我来说、这是一个瓶颈问题。 我 也可以分享测试代码。
您好、Ning、感谢您的答复。 VBUS - 5.23V
VBAT - 8.24V
03:13:41.839 -> ADC IBUS:0.01
03:13:41.839 -> ADC ICHG:0.00
03:13:41.839 -> ADC VBUS:5.16
03:13:41.839 -> ADC VBAT:8.34
03:13:41.839 -> ADC 顶部电池电压:3.97
03:13:41.839 -> ADC TS:83.79
03:13:41.839 -> ADC TDIE:22.50
03:13:41.839 -> ADC 电压底部单元:4.30
03:13:41.839 -> ICO 标志:1.
03:13:41.870 ->
03:13:41.870 ->充电状态:
03:13:41.870 -> IINDPM STAT:0
03:13:41.870 -> VINDPM STAT:0
03:13:41.870 -> TREG STAT:0
03:13:41.870 -> WD STAT:0
03:13:41.870 -> CHRG STAT:0
03:13:41.870 -> PG STAT:1.
03:13:41.870 -> VBUS STAT:1
03:13:41.870 -> ICO STAT:0
03:13:41.870 ->
03:13:41.870 -> NTC 统计信息:
03:13:41.870 -> NTC STAT:5
03:13:41.870 ->
03:13:41.870 ->故障统计信息:
03:13:41.870 -> VBUS OV 故障:0
03:13:41.870 -> TSHUT 故障:0
03:13:41.870 ->计时器故障:0
03:13:41.870 ->
03:13:41.870 ->电池平衡统计信息:
03:13:41.870 ->电池 BAL STAT:0
03:13:41.870 ->高侧 CV:0
03:13:41.870 ->低侧 CV:0
03:13:41.870 ->高侧 OV:0
03:13:41.870 ->低侧 OV:0
03:13:41.870 ->电芯 BAL OC:0
设置(我会使用这些设置)如果要查看代码设置配置、请让我知道
您好、Ning、感谢您的答复。 VBUS - 5.23V
VBAT - 8.24V
03:13:41.839 -> ADC IBUS:0.01
03:13:41.839 -> ADC ICHG:0.00
03:13:41.839 -> ADC VBUS:5.16
03:13:41.839 -> ADC VBAT:8.34
03:13:41.839 -> ADC 顶部电池电压:3.97
03:13:41.839 -> ADC TS:83.79
03:13:41.839 -> ADC TDIE:22.50
03:13:41.839 -> ADC 电压底部单元:4.30
03:13:41.839 -> ICO 标志:1.
03:13:41.870 ->
03:13:41.870 ->充电状态:
03:13:41.870 -> IINDPM STAT:0
03:13:41.870 -> VINDPM STAT:0
03:13:41.870 -> TREG STAT:0
03:13:41.870 -> WD STAT:0
03:13:41.870 -> CHRG STAT:0
03:13:41.870 -> PG STAT:1.
03:13:41.870 -> VBUS STAT:1
03:13:41.870 -> ICO STAT:0
03:13:41.870 ->
03:13:41.870 -> NTC 统计信息:
03:13:41.870 -> NTC STAT:5
03:13:41.870 ->
03:13:41.870 ->故障统计信息:
03:13:41.870 -> VBUS OV 故障:0
03:13:41.870 -> TSHUT 故障:0
03:13:41.870 ->计时器故障:0
03:13:41.870 ->
03:13:41.870 ->电池平衡统计信息:
03:13:41.870 ->电池 BAL STAT:0
03:13:41.870 ->高侧 CV:0
03:13:41.870 ->低侧 CV:0
03:13:41.870 ->高侧 OV:0
03:13:41.870 ->低侧 OV:0
03:13:41.870 ->电芯 BAL OC:0
设置(我会使用这些设置)如果要查看代码设置配置、请让我知道
IINDPM STAT:0
VINDPM STAT:0
TREG STAT:0
WD STAT:0
ICO STAT:1.
充电状态:未充电
电源正常状态:正常
VBUS 状态:适配器(3.0A)
------- NTC 统计数据-------------------------------------------------------
NTC 状态:冷
故障统计信息:
VBUS OV 故障:0
TSHUT 故障:0
定时器故障:0
电池平衡统计信息:
电池 BAL STAT:0
高侧 CV:0
低侧 CV:0
高侧 OV:0
低侧 OV:0
电池 BAL OC:0
ADC IBUS: 0.00
ADC ICHG:0.00
ADC VBUS:5.16
ADC VBAT:7.84
ADC CELL TOP:4.30
ADC CELL BOT:3.51
--------------------------------------------------------
这些是寄存器值、即使电池已放电、但仍然不在充电。
布局正确、因为该模块是 该供应商提供的 www.mikroe.com/balancer-5-click 标准模块、并且根据原理图、设计良好。 还有什么问题?